Chemical Stability and Compatibility - Heavy metal impurity testing in HNO3

 
Is it possible to prepare a mixture of these elements (Ni, As, Pb, Hg, Co, Cd) in one solution using 2% HNO3? Would there be any problems using ICP-MS with the same matrix?
Sort By    Oldest  |  Newest
Responses
admin Total posts: 529

The elements you list (Ni, As, Pb, Hg, Co, Cd) can be mixed together and are chemically stable in a 2% (v/v) HNO3 matrix with the following exception: Hg at concentrations typically used for ICP-MS analyses (<1 ppm) can suffer instability if the container used to store the solution is plastic (please see https://www.inorganicventures.com/mercury-chemical-stability for more information). As a result, for low concentrations of Hg in a 2% (v/v) HNO3 matrix it may be more appropriate to create and store the mixture in glass labware, though please be aware that any plastic that forms part of the ICP-MS sample introduction system (pump tubing, etc.) can also adsorb Hg and potentially create washout/memory effects.

Posted: 09/02/18 21:46:16