Measurement and Results - Poor lithium (Li) precision with new ICP-OES
| Responses | |
|---|---|
|
admin
Total posts: 529
|
If the precision of your Li measurements was satisfactory prior to using the new OES, and your samples have not changed significantly, then this would suggest that the new sample introduction system is the dominant variable affecting your repeatability. Generally speaking, precision variability between different emission lines would be a spectroscopic issue (e.g., due to interferences), and if the same poor precision is observed for multiple lines then the sample introduction system tends to be the likely culprit. If it’s possible to simply degas the samples by heating that would be a quick test of the influence of the dissolved hydrogen on the observed precision. Burgener Research offers a number of tips for optimizing Mira Mist nebulizers (http://www.burgener.com/MiraMistManual.html), and some of this advice might address the issues you are observing. As for the best Li line, we’d support a data-driven approach that includes careful examination of the spectra, comparison of axial vs. radial, and perhaps the evaluation of using an ionization buffer such as Cs to stabilize the ionization of Li. |