Question Responses Views Latest Reply
Sample preparation of cement using fused bead for ICP
in ur application XRF VS ICP for concrete metals u have given a sample preparation using fused bead and then dissolving in Dilute nitric acid, i wanted to know whether the sample dissolves completely in the dil nitric acid or do you get some insoluble for that and then u filter to inject in ICP ?
Created: 11/13/19 04:09:38
Created By: "Lucia Braganza"
2 3749
Sample preparation of cement using fused bead for ICP
in ur application XRF VS ICP for concrete metals u have given a sample preparation using fused bead and then dissolving in Dilute nitric acid, i wanted to know whether the sample dissolves completely in the dil nitric acid or do you get some insoluble for that and then u filter to inject in ICP ?
Posted: 11/13/19 14:23:18
Struggling with Analyte recory and accuracy for Primary lead impurities by ICP-OES
I am currently trying to develop a method for determining impurities in primary lead and recycled secondary lead. I have been struggling to get good recovery for my elements the current method of digestion used involves using Nitric Acid, perchloric acid and precipitating out of lead using hydrochoric Acid. I used a CRM as a sample and i was not getting the expected concentrations for my Elements. I dont
Created: 09/26/19 06:29:33
Created By: "Simphiwe Khambule"
2 3299
Struggling with Analyte recory and accuracy for Primary lead impurities by ICP-OES
I am currently trying to develop a method for determining impurities in primary lead and recycled secondary lead. I have been struggling to get good recovery for my elements the current method of digestion used involves using Nitric Acid, perchloric acid and precipitating out of lead using hydrochoric Acid. I used a CRM as a sample and i was not getting the expected concentrations for my Elements. I dont
Posted: 11/12/19 14:23:56
Dissolve PtPdRu alloy
I have some catalyst composed of PtPdRu/C. However, Pd and Ru are difficult to get dissolved even in aqua regia. Is there any solution to get their composition from ICP?
Created: 08/09/19 10:33:34
Created By: "longsheng cao"
1 3079
Dissolve PtPdRu alloy
I have some catalyst composed of PtPdRu/C. However, Pd and Ru are difficult to get dissolved even in aqua regia. Is there any solution to get their composition from ICP?
Posted: 08/19/19 11:07:25
Sample Recovery- Open-cup evaporation

Hi all,

We recently purchased a ThermoFisher iCap RQ ICP-MS. We are analyzing 14 metals in our Dimethyl Sulfoxide (DMSO) product. We utilize open-cup evaporation on a hot plate followed by reconstituting in 2%HNO3 & 0.5% HCl. Currently, I am adding 30mL of DMSO sample into a teflon beaker, then spiking from a stock solution an aliquot equivalent to 1ppb of the 14 metal solution. I allow the sample to evaporate to dryness (between a pin-tip and a light char mark). I then use my diluent to squirt into the cup ~5-10mL. This portion is transferred to a sample vial and diluted back up to 30mL. 

I am struggling with recovery. The data is inconsistent to say the least. My 2ppb samples are sometimes lower than my 1ppb samples. Sometimes they are 2-3 times higher than they should be. 

Performance reports look great, and standard curve looks great. 

Any tips for improving recovery of metals following evaporation? Any help would be highly appreciated. Thanks in advance.


Andrew

Created: 08/08/19 08:42:18
Created By: "admin"
1 2649
Sample Recovery- Open-cup evaporation

Hi all,

We recently purchased a ThermoFisher iCap RQ ICP-MS. We are analyzing 14 metals in our Dimethyl Sulfoxide (DMSO) product. We utilize open-cup evaporation on a hot plate followed by reconstituting in 2%HNO3 & 0.5% HCl. Currently, I am adding 30mL of DMSO sample into a teflon beaker, then spiking from a stock solution an aliquot equivalent to 1ppb of the 14 metal solution. I allow the sample to evaporate to dryness (between a pin-tip and a light char mark). I then use my diluent to squirt into the cup ~5-10mL. This portion is transferred to a sample vial and diluted back up to 30mL. 

I am struggling with recovery. The data is inconsistent to say the least. My 2ppb samples are sometimes lower than my 1ppb samples. Sometimes they are 2-3 times higher than they should be. 

Performance reports look great, and standard curve looks great. 

Any tips for improving recovery of metals following evaporation? Any help would be highly appreciated. Thanks in advance.


Andrew

Posted: 08/19/19 11:23:50
Testing
More testing
Created: 07/12/19 15:11:00
Created By: "longsheng cao"
0 705
No Data Found
Testing
This question is a test.
Created: 07/12/19 12:27:15
Created By: "longsheng cao"
0 785
No Data Found
Sample preparation without organic solvents
I am interested in measure the Pb to S ratio in PbS nanocrystals (NC) using an ICP-OES instrument. These NC are dispersed in an organic solvent and stabilized with long chain ligands like oleic acid. I can also deposit the NC as a film to remove the solvent.My concern is that when I digest the PbS NC film the acid will volatilize the sulfur thus skewing my results. Would the universal solvents you use for zeolite measurements work on such a system? Any other sample prep advice for this system is most appreciated. FYI: Everything entering our ICP-OES has to be in an aqueous solution, we do not have an organic solvent line.
Created: 06/14/19 10:17:35
Created By: "admin"
1 2729
Sample preparation without organic solvents
I am interested in measure the Pb to S ratio in PbS nanocrystals (NC) using an ICP-OES instrument. These NC are dispersed in an organic solvent and stabilized with long chain ligands like oleic acid. I can also deposit the NC as a film to remove the solvent.My concern is that when I digest the PbS NC film the acid will volatilize the sulfur thus skewing my results. Would the universal solvents you use for zeolite measurements work on such a system? Any other sample prep advice for this system is most appreciated. FYI: Everything entering our ICP-OES has to be in an aqueous solution, we do not have an organic solvent line.
Posted: 06/14/19 10:18:04
Low-alloy steel analysis by ICP
We need to perform low-alloy steel analysis by ICP, and wanted to know what would be considered an adequate sample size in grams. Our sample digestion uses acids under pressure in a microwave oven.
Created: 06/14/19 10:15:46
Created By: "admin"
1 2550
Low-alloy steel analysis by ICP
We need to perform low-alloy steel analysis by ICP, and wanted to know what would be considered an adequate sample size in grams. Our sample digestion uses acids under pressure in a microwave oven.
Posted: 06/14/19 10:16:57
Digesting Pt oxide
I'm trying to dissolve Pt isotopic spikes. I've dissolved Pt metal before with no problem by first dissolving the metal in concentrated aqua regia in a teflon beaker, then evaporating the aqua regia by placing the beaker on a hotplate with the lid off, redissolving the residue in concentrated HCl, evaporating that, and ultimately dissolving this residue in 6M HCl. The Pt isotopic spikes are in the form of a very fine powder. In fact the powder was so fine that I could not remove all of it from the vials without washing them out with something. I chose to wash the vials out with methanol because I could evaporate it quickly afterwards both to begin the aqua regia dissolution and to reweigh the shipping vial after removal of the spike so that I knew how much spike I had dissolved. I also thought methanol would not react with the Pt at low temperatures. After washing the Pt+methanol into teflon beakers, I put the beakers (unlidded) on a hotplate set to 100 C to evaporate. After they were totally dry, I added the concentrated aqua regia and allowed the spikes to dissolve for two days. After two days, the spikes appeared to be dissolved, although it was hard to tell since the powder is so fine. I have dissolved a greater mass of Pt wire in the same amount of time, so I suspect the spikes were dissolved, or would have dissolved if they were still Pt metal. I evaporated the aqua regia and instead of the normal dark red residue (which dissolves in conc. HCl in seconds), I had a black residue. I added the concentrated HCl anyway, and a small amount of residue dissolved but most did not, even after sitting in lidded beakers on a hotplate for 24 hours. This has happened with two Pt metal spikes that were on a hotplate at 100 C with methanol and never with Pt metal that did not contact methanol, so it seems that Pt and methanol do react at temperatures as low as 100 C.I've tried re-doing the aqua regia step and have gotten to the point where I now have both an easily-dissolved reddish residue and an insoluble black residue in each beaker. I need to dissolve everything because I don't want to fractionate my isotopic spikes. But the problem is I have no idea what I made and so no clue how to ultimately get it to dissolve into 6M HCl. Do you have any suggestions?
Created: 06/14/19 10:14:03
Created By: "admin"
1 2779
Digesting Pt oxide
I'm trying to dissolve Pt isotopic spikes. I've dissolved Pt metal before with no problem by first dissolving the metal in concentrated aqua regia in a teflon beaker, then evaporating the aqua regia by placing the beaker on a hotplate with the lid off, redissolving the residue in concentrated HCl, evaporating that, and ultimately dissolving this residue in 6M HCl. The Pt isotopic spikes are in the form of a very fine powder. In fact the powder was so fine that I could not remove all of it from the vials without washing them out with something. I chose to wash the vials out with methanol because I could evaporate it quickly afterwards both to begin the aqua regia dissolution and to reweigh the shipping vial after removal of the spike so that I knew how much spike I had dissolved. I also thought methanol would not react with the Pt at low temperatures. After washing the Pt+methanol into teflon beakers, I put the beakers (unlidded) on a hotplate set to 100 C to evaporate. After they were totally dry, I added the concentrated aqua regia and allowed the spikes to dissolve for two days. After two days, the spikes appeared to be dissolved, although it was hard to tell since the powder is so fine. I have dissolved a greater mass of Pt wire in the same amount of time, so I suspect the spikes were dissolved, or would have dissolved if they were still Pt metal. I evaporated the aqua regia and instead of the normal dark red residue (which dissolves in conc. HCl in seconds), I had a black residue. I added the concentrated HCl anyway, and a small amount of residue dissolved but most did not, even after sitting in lidded beakers on a hotplate for 24 hours. This has happened with two Pt metal spikes that were on a hotplate at 100 C with methanol and never with Pt metal that did not contact methanol, so it seems that Pt and methanol do react at temperatures as low as 100 C.I've tried re-doing the aqua regia step and have gotten to the point where I now have both an easily-dissolved reddish residue and an insoluble black residue in each beaker. I need to dissolve everything because I don't want to fractionate my isotopic spikes. But the problem is I have no idea what I made and so no clue how to ultimately get it to dissolve into 6M HCl. Do you have any suggestions?
Posted: 06/14/19 10:14:34
Neutralizing HF with TEA (triethanolamine)
A question has come up regarding the neutralization of HF using TEA (triethanolamine).The digested sample will have residual HF, and we are asking what concentration of TEA may need to be added to perhaps 1 to 2% residual HF. The elements of interest in the sample are Mg and Ti.
Created: 06/14/19 10:11:46
Created By: "admin"
2 2501
Neutralizing HF with TEA (triethanolamine)
A question has come up regarding the neutralization of HF using TEA (triethanolamine).The digested sample will have residual HF, and we are asking what concentration of TEA may need to be added to perhaps 1 to 2% residual HF. The elements of interest in the sample are Mg and Ti.
Posted: 06/14/19 10:12:21